Using a Disc tip / probe
It is common to use a "disc" stylus when inspecting parts which contain an "undercut" such as a snap ring groove.  Here are some tips on proper usage of disc tips in CMM-Manager.
Alert
For many technical reasons as well as hardware and calibration requirements outside our control, inferred calibration / head touches with PH20 and Revo are not supported.
When using an I++ DME configuration (normally SP25 or PH20/REVO), the probe configuration and funcitonal support is as specified in UCC Server.   This article does not (directly) apply to I++ configurations.
Configurations running Direct Connect 2.0 will behave as described here.
Support
CMM-Manager supports the use of disc probes for both touch (TP20) and analog (SP25) scanning.
There are two types of disc probes. Flat faced (cylindrical slice) and round faced (spherical slice).  CMM-Manager supports the spherical slice (typically ruby) probe.  The spherical slice is much more common and the preferred choice for reasons noted below in 'Why a spherical slice is better than a cylindrical slice'.  If you wish to use a cylindrical slice disc we suggest you define it as a cylinder tip.  Due to the fact that a disc probe is a very thin slice, the usage of the cylindrical face tip will be very limited.
Many disc probes contain a pair of "nubs" that could be calibrated and used for width measurements.  The nubs are not supported in CMM-Manager.  CMM-Manager does not support using the disc along the probe axis for measuring  a width with or without the nubs.  The nubs require a special calibration procedure and many disc tips allow "slipping" the 'tip' around the shank to orient the nubs.  This means that the position is not repeatable and several probe indexes would be required to acquire points using many calibrated angles for both the upper and lower nub.
In the real world, due to manufacturing tolerances and build stack-up errors, a disc probe will never be exactly square to the surface to be inspected.  This limits the accuracy of cylindrical disc tips during calibration as well as during use.  Although not supported, it would limit the accuracy of any disc tip used without the nubs for a width measurement.
For calibration, the "square" examples below demonstrate the issue with calibrating a cylindrical disc on a ring gauge surface, as well as using it for inspections.  The sphere examples demonstrate the issue calibrating the cylindrical disc on a sphere or using it for inspection of curved surfaces.
The following examples visualize the problems.
Defining a disc probe
Note:  The first probe assembly (Probe Assembly 0 by default) cannot be a disc probe
    If you select 'Disc' as the type in CMM-Manager, you will be prompted to enter the dimensions for a simple disc tip.   You will be asked to enter the dimension from the shank face at the thread to the bottom of the disc, the disc diameter, as well as the disc thickness.
Calibrating disc probes
    When calibrating a disc probe on a sphere in CMM-Manager, the calibration path will use two circular paths, one slightly above and one slightly below the 'equator'.  For instance, a 3mm thick disc may calibrate roughly 1mm above and 1mm below the equator.  This stops the touches from "breaking over" the edge of the spherical surface to the flat face(s).
When CMM-Manager calibrates a disc for scanning, several touches are taken above and below the sphere "equator" to determine the relationship of the tip and the sphere.  An optimize scan path is then created that works as above.
    A ring gauge may also be use for calibration of disc probes for touch (TP20).  When calibrated using a ring gauge, the tip will only be accurate when used to inspect features that are in the plane of the disc during calibration.  No "Z" axis compensation is calculated.
Note: The ring gauge must be defined and located before use, just like a calibration sphere.
Using a Disc probe
Disc probes may only be used for inspections where the surface(s) to be inspected are aligned with the face of the disc.  Many probe heads will only index in 7.5 degree increments severely limiting the angle of the surfaces that can be inspected.  No check is performed by CMM-Manager as to the validity of the inspection on angled surfaces.