CMM-Manager supports Plane measurement by using touches or scanning.  Multiple methods are available for each, depending on the CMM probing hardware.
Creating a Scan Inspection
Due to the nature of the hardware during scanning it is sometimes necessary to control the head motion differently than the tip scanning motion.  Instructions for controlling the head motion during a scan can be found in Controlling Head Motion during 5-Axis Scanning.
3D Scan
Probing Support: PH10 (SP25), REVO (RSP2 & RSP3)
Alert
Please note that RSP2 for REVO is not designed for scanning perpendicular to a surface. Therefore, the user may need to select an appropriate tip prior to the measurement or utilize RSP3.
Default: Parallel Lines
The default 3D Scan Method creates a series of scan lines back and forth at even intervals across the plane surface.  The ends of each scan line are connected to the end of the next in a rectangular back-and-forth scan pattern.
  1. Define or select a plane from CAD to be measured.
  2. Select the 3D option in the Scan Method frame.
  3. Select the # of Scan Lines to be added from the Scan Flow frame as well as the Margin to be used as an offset from the edges of the plane.
    1. You may change the direction of the main scan lines by selecting or de-selecting the Along Side 1 checkbox.
  4. Click Create Path to update your scan path.
Alert
Path planning does not check for voids in the surface for 3D scan paths.  It is possible to edit the end points for each scan line on the Points Tab to shift the line to clear obstacles.
Manually Picked Points
  1. Points may be added manually by using the tools available on the Points Tab.  Points may be entered manually by using the Keyin Point button or selected directly from CAD by using the Add Pick tool.
  2. It is possible to create a path using only points selected by the programmer.
    1. Use Select All to select all the points in the dialog.
    2. Use Delete Selected to remove all of the points.
    3. Use either Pick Points or Keyin Points to enter the desired points for the inspection.
Be sure to click Create Path to apply your settings.
Gasket Scan
Probing Support: PH10 (SP25), REVO (RSP2 / RSP3)
The Gasket scan allows very complex scan paths to be created.  Gasket scans can be created using 3-axis motion or can be created using 5-axis motion with a REVO.  5-axis motion scans include several options for controlling how probe head motion is utilized while following the scan path that the tip follows.
Default Gasket Scan
The default gasket scan creates a series of scan lines back and forth at even intervals across the plane surface just like the 3D scan.  The ends of each scan line are connected to the end of the next in a rectangular back-and-forth scan pattern.  The Gasket scan also includes several options for controlling the head motion while the tip follows that scan path.
  2. Select the Gasket option in the Scan Method frame.
Alert
Path planning does not check for voids in the surface for Gasket scan paths.  It is possible to edit the end points for each scan line on the Points Tab to shift the line to clear obstacles.
Complex Gasket Scan
The user may define a complex gasket scan by using the Points Tab and entering a series of points that the probe will follow.  If the Override DCC Settings option for Fly Radius is turned on, the head motion will be "blended" by the controller when the scan is executed.
  3. Select the Points Tab.
  4. Click Select All and Delete to remove any existing points.
  5. Click Pick Points and select points on CAD to create the desired path.  Use extra points around curves to "smooth" the path during execution.
    1. See Controlling 5-Axis Head Motion above to change how the head motion is controlled during the scan.
  6. Click Create Path to update your scan path.
Sweep Scan
Probing Support: REVO (RSP2)
Sweep scans are a method of collecting large amounts of data by rapidly sweeping the tip back and forth while moving the probe along a defined path.
Simple Sweep Scan
The simplest Sweep scan is a scan between a start and an end point.  The probe head will move along the path between the start and end point while the tip sweeps back and forth rapidly at the same time.  The area from which data is collected is typically a simple rectangular area.
  2. Select the Sweep option in the Scan Method frame.
  3. A default path (start and end point) will be created.
    1. The probe head will move along Side 1 and the tip will sweep back and forth along the direction indicated by Side 2.  The direction of the scan can be changed by unchecking Along Side 1.
    2. The default # of Sections is 2.  Each section defines a start point, end point, or way point.  The minimum number of sections is 2.
    3. The width of the sweep may be adjusted by changing Width.
  4. In order to control the amount of data collected and the number of "sweeps" between the start and end point, check the Overwrite Scan Param. checkbox.  Then select the Scan Tab.
    1. Pitch determines the spacing between collected points in current units.
    2. Line Pitch determines the distance between each sweep crossing of the head path in current units.
  5. Click Create Path to update your scan path.
Complex Sweep Scan
The user can create complex sweep scans by selecting the Points Tab and entering a series of sections (automatically defined by user point selection).  The motion between the sections will be smoothly blended and the width of the sweep will be dynamically adjusted along the path to meet the requirements for each segment.
  3. In order to control the amount of data collected and the number of "sweeps" between the start and end point, check the Overwrite Scan Param. checkbox.  Then select the Scan Tab.
  4. Select the Points Tab.
  5. Click Select All and Delete to remove any existing points.
  6. Click Pick Points and select points on CAD to create the desired path.  Use extra points around curves to "smooth" the path during execution.
    1. CMM-Manager will automatically adjust the orientation of the series of sections to create a smooth sweep path.
    2. After the points (sections) have been created, double click individual points, or click a point and then click Modify in order to change the width of the scan at that section.  The width from the previous section and to the following section will be dynamically blended during execution.
  7. Click Create Path to update your scan path.