When using a Renshaw UCC controller with UCC Server 5.0 and above in I++ mode, sync of these values should be automatic. If the settings are still not correct, beginning in CMM-Manager 2022, the preferred method is to use the instructions here
. If you are using a system with a UCC Server version older than 5.0, please see instead:
Setting table height with UCC Server I++Beginning with CMM-Manager 2020, a new feature was added to the machine volume display when turned on via Display | Machine Volume. It is a second bounding line that should be just above the table.
The purpose of displaying this boundary is to indicate the lowest that the center of rotation of the probe head can go. This keeps the probe head from colliding with the table surface when a tip angle is selected where the tip is higher than the bottom of the probe head.
This does not affect machine limits when a probe angle is selected where the tip is "below" the bottom of the probe head. You can still take touches on the table surface in that case.
While this has always been the case, it was sometimes not clear why a "collision with machine boundaries" was displayed.
Note: It is sometimes possible for the probe tip to rotate outside the machine limits while selecting a new probe angle. You cannot, however, use any kind of command that creates motion outside the machine limits.
Setting the Machine Volume
- Select a probe angle such that the probe head can be moved to the machine extreme limits while the tip will not collide with any machine components. Any calibrated or non-calibrated probe may be used.
- Select System | Machine Setup from the Ribbon Bar. The Machine Setup dialog will be displayed (The password is service). Select the Machine
tab. - For UCC I++ systems, an Sync with I++ Server
checkbox should be un-checked before continuing.
- Move the machine to the extreme X axis Min and Max limits. Click the appropriate [<] button in the Machine Setup dialog to set the values. Be sure you pay attention and place select the correct value (Min/Max).
- Repeat this for the Y and Z axes.
NOTE: For the Z axis Min, the bottom of the probe head should typically be just above the table. If you wish a larger safety margin, you may leave a larger gap. For a PH10, this bottom boundary is ~ 25mm above the top of the table (distance from the top of the table to the center of rotation of the PH10). You may also desire to use a permanent fixture plate as your lower limit instead of the table.
As a shortcut, you may move the machine to the two extreme opposite corners. Please note that any particular corner may not be Min or Max for all 3 axes!
Click Apply and OK to close the Machine Setup dialog.
Setting the Table Height
- Select a probe angle that can touch the table, usually A0B0 for a bridge machine. Using the calibrated Ref Tip or another calibrated tip is recommended.
- Select MORE... from the Status Bar (far right end). This will open the Machine Property dialog.
- Select Table | Update Table Position.
- As instructed, move the tip as close to the table as possible, then click OK to complete the setup.