ACR3
The Renishaw ACR3 tool change rack is used with Renishaw probing systems, which utilize the AutoJoint style tool connection system. The ACR3 Tool Changer performs the same function as the older style ACR1 rack, but the ACR3 is passive, meaning no electrical power or control signals are used. The ACR3 is a module that mounts to the Renishaw Modular Rack System.
Installing the ACR3 Rack
The ACR3 rack is installed as part of the Renishaw Modular Rack System. The ACR3 module is attached to the Modular Rack with 'T' nuts and screws. The Modular Rack System may be mounted in any convenient location in the machine volume. The rack must be aligned with the machine X or Y axes. Renishaw recommends that the rack is within 100 microns of being square to the appropriate axis.
Detailed instructions may be found in the ACR3 auto change rack system Installation and user's guide, document H-1000-5087-01-A, available from Renishaw.
Configuring the ACR3 Rack
Note: If you are using an ACR3 rack with a system using the I++ protocol, the configuration is done in the I++ Server software.
The following steps are required to configure the ACR3 rack for use.
Making the ACR3 available to CMM-Manager
The first step is to let CMM-Manager know that a rack has been installed. This is done from System > Setup > Machine Setup on the Ribbon bar.
Select the Probe tab from the Machine Setup dialog, select the type of rack you have installed from the drop-down list, and click ADD to add the rack to the list of available racks. More than one rack may be installed at one time. Depending on the probe configuration, various styles and quantities of racks may be mixed and matched.
Once you have added any racks to the Machine Setup, click Apply to save the changes, then click OK to close the dialog.
Setting the ACR3 Parameters
Note: CMM-Manager must be ONLINE for the Configure dialog to become available.
The rack configuration may be edited by selecting the Probe tab > Rack group > Configure. If multiple racks are installed, select the correct one from the Station Index drop-down list.
Reasonable default settings are supplied as a starting point for Station Speeds. Speeds are in current units (mm/sec or in/sec).
By default, the Use Safe Pos Btw Ports is NOT enabled. In the normal procedure, a rack will be approached from the safe position, then the first tool placed into the rack, and finally, a simple move across the front of the rack to the pickup location will be performed. If this check box is checked, the machine will move to the safe position before moving to the next port. This only applies to movement between ports on one rack. Changes that access multiple racks will still move to the safe position between each rack.
The remaining values are filled in when the rack position is taught.
Assigning Components to a Port (ACR3)
Next, we need to assign the port to use to store the various probe assemblies. Not all ports must be filled, and not all probe assemblies must be assigned a port.
Note: If a probe assembly is not assigned to a port and that tool is requested during a program, the user will be prompted to place that probe assembly on the probe head manually.
After creating definitions for all of the available tools using the Probe Assembly dialog, select each tool, in turn, click assign, then select the probe station and the port that will be used to store that probe assembly. Click OK on the Assign Station and Port dialog, then continue through the rest of the tools in the same manner.
Note: This example is a basic example indicating how to store the entire probe assembly into one port. For examples of complex tool changes, see Multiple Rack Configuration.
Teaching the ACR3 Rack Location
Note: Before attempting this procedure, you must have added an ACR3 Probe Station in Machine Setup. The Modular Rack System must be squared to the machine axes. (Renishaw suggests within 100 microns)
- Move the ACR3 rack to the unlock (left) position.
- Insert lid clips and the setting gauge as shown.
- Calibrate and select a tip at the A and B angles required to allow the probe to be stored in the ports in the ACR3.
Due to requirements later in this procedure, you MUST use the angle that is used for docking.
- Select Probe Tab > Rack Group > Calibrate in the Ribbon Bar (or click the Calibrate Station button in the Probe Station dialog)
- Select the ACR3 probe rack to calibrate from the dropdown list and click OK.
- Verify that steps 1 - 3 listed in the Things to Do to have been completed.
- Select the appropriate parameters in the ACR3 Port Configuration section.
- Ports #: The number of ports (4 or 8)
- Direction (Unlock->Lock): The axis and direction of motion to move from the Unlock to the Lock position (ACR3 moving to the right)
- Click OK.
- Perform the following steps in the Jog in Machine Ref. Frame section to jog the probe into the appropriate Z position.
- Select the machine reference frame axis from the dropdown.
- Type in the amount of motion (in the current measurement units) for a jog movement.
- Click the >> button to jog in the positive direction of the selected axis (or << to move in the negative direction).
Proper position: - The latching tip should be centered L/R in the guide.
- The body should be near the back of the port as though it was in position for picking up a component.
- The face of the PAA pickup should "just touch" the top face of the guide. There is about 0.5mm "play" up and down when the guide is in the slot. Some systems "calibrate" differently. It may be necessary to repeat the process and hold the guide in the UP position in the slot while "just touching" it with the face of the PAA pickup.
- Click Get Position when the probe is in the correct position.
Proper position: - The LARGE diameter of the PAA pickup, NOT the smaller diameter step, should "just touch" the guide. Too much force will cause the head to unlock.
- Move the probe head until it is clear of the ACR3, connect the probe assembly, remove the setting gage, and click OK.
- Measure four points on the top plane of Port 1.
Measure four points in the back arc of Port 1. Be sure to avoid the lock and mounting holes.
Note: CMM-Manager will switch to DCC and measure the remaining points. You will be prompted to move the rack from the Unlock to the Lock position during this procedure.
- Once all ports have been measured in both the lock and unlock position, the user will be prompted to manually move the ACR3 to the left so port 4 lines up precisely with the probe's location. Slide the ACR3 to the left until it visually lines up behind the probe assembly. There is a small amount of "spring back" in the ACR3, so you may need to move slightly beyond the required position, so it settles in the proper location.
- Click OK. The probe will VERY slowly (1mm / sec) dock itself in port 4.
- Using slow speed on the joystick, and being VERY SURE not to move in the Z or docking direction accidentally, move the rack left/right until the circle is exactly centered in the window.
Click OK.
The probe will automatically move clear of the rack.
Move the probe to a safe position for movement to and from the rack.
The safe position must account for any probe rotation that will occur at this position. During a probe change, the probe will move to this position with the current A/B angles then rotate after reaching the safe position. To reduce the amount of free space required around the safe position, program a series of clearance moves and a probe rotation before requesting a probe change. Be sure to allow enough clearance in any case in front of the rack for any star or "crank" styli.
- Click OK.
This completes the rack location procedure. The first few tool changes should be done at slow speed to verify that the rack has been properly configured.
Loading Components into the ACR3 Rack
Before the rack is loaded with tools, it must first be configured properly, and it's location taught.
- Once the proper setup is complete, we can begin loading the rack.
- Begin by manually assembling the first probe assembly and mounting it to the probe head.
- Select Probe > Probe Assembly, select the tool currently on the probe head and click the "Set Active" button. This will synchronize CMM-Manager with the actual hardware.
- Select Probe > Probe group > Get Probe Assembly and select the next tool. CMM-Manager will place the tip into the rack and prompt the operator to place the next component on the current tool manually.
- Continue the Get Probe Assembly process until all tips have been placed in the rack. Be sure to return to the original probe assembly as the last step to complete the rack loading.
- Be sure to do this for all racks when multiple racks are used.
To use the ACR3, from the Ribbon Bar, select Probe > Probe group > Get Probe Assembly. Select the desired probe from the drop-down list and click OK. The machine will then go and automatically load the proper tool.
Any other Get Probe Assembly commands used in a program will cause the machine to do a tool change automatically.
Warning: When writing a program, be sure there is a clear path to the probe rack when selecting a new tool. It may be necessary for the programmer to add their own clearance moves to ensure a collision-free tool change.
Using the Probe Stations configure dialog, various ACR3 information can be modified.
Once the rack location has been taught, the Location, Safe Pos., Front Dir, and Up Dir. fields will be filled in. Of the fields filled in by the teaching process, only the Safe Pos can be edited.
If it is noted that all ports of the rack seem a little high or a little low, causing interference during a change, an overall offset can be applied. The value entered into the Offset field will be added to the value listed in the Location field. To move ALL ports up 100 microns (with units set to mm), type .100 (note the leading decimal point) into the Z cell of the Offset row. To move down, enter a negative value (-0.100)
If only ONE port needs to be modified, entering an offset in the appropriate cell for an individual port will adjust just that one port by the value specified.
If the rack has been properly taught, It is exceptionally rare that offsets would have to be applied.
If a large probe is used and needs a greater approach and back-off from the rack, enter the offset in the station's Large Probe Offset field.