Tip Manager is the primary control window used to define, calibrate, inspect, and manage all probe tip orientations (A and B angular positions on indexable heads like a PH10, as well as stylus definitions) for your active probe assembly. From this window, you can...
Key Functions of Tip Manager
- Creating Tips and Tip Sets
- New Tip: Adds a single tip angle by specifying exact A and B angles, nominal stylus ball diameter, and stylus index (in the case of a star stylus).
- New Set: Automatically generates a batch set of tips at user-defined A and B angular increments (e.g., every 30° or 90°)
- Tip Qualification & Calibration
- Calibrate: Qualifies a single selected tip against a reference sphere or ring gauge to calculate its precise X, Y, Z center offset vector and ball diameter.
- Calibrate All: Initiates automated batch calibration for all non-calibrated or program-specific tips in one sequence.
- Viewing Tip Information
- View Details: Displays detailed metadata for any highlighted tip, including calculated X, Y, Z center offsets relative to the reference tip (_RefTip), stylus ball diameter, form error from the last qualification run, and calibration timestamp.
- Selecting the Active Tip
- Select: Sets the highlighted tip as the active inspection tip. In Teach-On Mode, clicking Select inserts a Get Tip command into your Program Database and updates the 3D graphic display.
- File Management
- Save / Load: Saves probe tip lists to .tip resource files or loads pre-configured tip files into your configuration for import/export purposes.
How to Access Tip Manager
You can open the Tip Manager window using two methods:
- Ribbon Bar: Go to Probe Tab > Tip Group > Tip Manager.
- Status Bar: Click directly on the Active Tip pane in the Status Bar at the bottom of the CMM-Manager window, which displays the name of the currently active tip.
Creating New Tips
New Tip
This function allows you to create a custom probe tip orientation by specifying the following parameters:
- Name: Accept CMM-Manager's automatically assigned default name (such as A45B90) or type a custom, unique label for the tip.
- A Angle & B Angle: Enter the desired rotation angles directly into the text fields, or use the scroll bar below each field to adjust the angles according to your probe head's increment resolution. CMM-Manager will automatically update the tip name based on the specified A and B angles.
- Diameter: Type the nominal stylus ball diameter in your active project units (mm or inches). This field should be auto-populated and correct if the probe is defined correctly in the Probe Assembly window.
- Stylus Index: If your active probe head uses a star stylus, select the appropriate tip branch index from the drop-down list.
- Stylus 0: The vertical stylus mounted directly along the contline of the touch probe.
- Stylus 1: The horizontal branch pointing toward the front of the machine (i.e. same direction as the probe head LED).
- Stylus 2: The horizontal branch pointing toward the right side of the machine.
- Stylus 3: The horizontal branch pointing toward the back of the machine.
- Stylus 4: The horizontal branch pointing toward the left side of the machine.
The X, Y, and Z Offset vector fields are automatically calculated by CMM-Manager relative to the reference tip (_RefTip) based on the specified A and B angles. Therefore, these values should be correct and should not need to be modified, provided that the probe assembly is defined correctly.
New Set
Use the New Set button to create a batch of probe tips at specified angular increments. This feature is particularly useful when setting up a new probe assembly and quickly creating multiple tip orientations for initial calibration and programming.
Tip Set Parameters are shown below:
- Tip Prefix / Tip Suffix: Enter optional text to append at the start or end of each generated tip name (for example, entering Tool1_ as a prefix creates tip names like Tool1_A0B0, Tool1_A30B90, etc.)
- A Interval: Set the angular step size for the A angle (vertical tilt) by moving the slider or typing a value.
- Fix A: Check this box if you want to lock the A angle at a specific fixed value (e.g., fixing A at 90.0°) rather than stepping through a range.
- B Interval: Set the angular step size for the B angle (horizontal rotation).
- Fix B: Check this box to lock the B angle at a fixed angle.
- Stylus Index: Select the stylus index from the drop-down menu if using a star probe assembly.
For example, let's assume that following parameters were entered in the New Set window for a machine equipped with PH10 probe head.
- Tip Prefix: Tool1_
- Tip Suffix: (left blank)
- A Interval: 30.0° (Fix A: Unchecked)
- B Interval: 90.0° (Fix B: Unchecked)
Since the range of PH10's A-axis is 0° to 105° and B-axis is ±180°, 16 total tips will be generated.
A Angle | Generated Tip Names |
A = 0° (1 tip) | Tool1_A0B0
A = 30° (5 tips) | Tool1_A30B0, Tool1_A30B90, Tool1_A30B180, Tool1_A30B-90, Tool1_A30B-180
A = 60° (5 tips) | Tool1_A60B0, Tool1_A60B90, Tool1_A60B180, Tool1_A60B-90, Tool1_A60B-180
A = 90° (5 tips) | Tool1_A90B0, Tool1_A90B90, Tool1_A90B180, Tool1_A90B-90, Tool1_A90B-180
Tip calibration
Calibrate
Use the Calibrate button to qualify the selected probe angle against a located reference artifact (such as a calibration sphere or ring gauge).
To calibrate a desired tip...
- In Tip Manager, click to highlight the specific tip angle (e.g., A0B0, A45B90) you want to calibrate.
- On the right-hand action panel of the Tip Manager window, click Calibrate. The Calibrate Tip dialog box will pop up.
- In the Calibrate Tip dialog box, adjust the setup fields as needed.
- Cal. Sphere: Select the desired located calibration sphere from the drop-down list (e.g., Def. Cal. Sphere)
- For disc or cylinder styli, you can also select Cal. Ring if using a ring gauge.
- DCC: Keep the DCC checkbox checked to run the calibration automatically under machine control, or uncheck it to measure manually using the joystick.
- Calibration Point #: Set the number of touch points to collect on the sphere (e.g., default 5 points for standard touch-trigger probes)
- Scanning Parameters (SP25): If using an analog scanning probe, set the Calibration Speed, Probe Deflection 1 & 2, or check Quick Calibration.
- Click OK (or Calibrate)
- A Clear Probe Path prompt will appear; ensure the CMM path between the probe and calibration sphere is free of obstructions, then click OK.
- You'll be prompted to manually move the probe tip above the calibration sphere instead if you are calibrating a reference tip (_RefTip).
- In DCC mode, the head automatically indexes to the chosen A & B angles and sweeps/touches the sphere. Take the required touch points manually if you unchecked the DCC option.
- Once probing is finished, a flyby result box displays the calculated Tip Diameter and Form Error.
- In the Tip Manager, the tip status indicator turns Green (showing it is calibrated), and the updated calculated Tip Diameter and Form Error are recorded in the table.
Calibrate All (Batch Tip Calibration)
Calibrate All should be used instead of Calibrate if you would like to calibrate more than a single tip as this enables you to calibrate multiple tips based on a given rule.
When the Batch Tip Calibration dialog opens after clicking the Calibrate All button, set the following options according to your requirements.
- Calibration Scope
- All non-calibrated tips: Qualifies only those tip angles in the list that currently lack calibration (i.e. gray calibration frag).
- All tips: Forces re-calibration of every tip angle in the Tip Manager, regardless of current calibration status.
- Only non-calibrated tips in current program: Limits calibration strictly to uncalibrated tips required by the currently opened part program.
- All tips in current program: Re-calibrates every tip needed for the active part program while skipping unused tips.
- Calibration Point #: Sets the number of touch points taken per tip on the calibration sphere (e.g., 5 points).
- Calibration sphere: Select All or choose a specific located calibration sphere from the drop-down list.
- Prefer default calibration sphere: Instructs CMM-Manager to use the default sphere if applicable.
- Save tip summary: Saves a text report (<ProgramName>_TipsSummary.txt) to CMM-Manager's Resource folder containing final tip diameters and form errors upon completion for traceability.
- Add to program: Inserts the batch calibration command into the active part program so it executes automatically during future program runs.
- Establish Ref. Tip: Forces the routine to first re-qualify the reference tip (_RefTip) on the sphere before proceeding to calibrate remaining tips
- All Probe Assemblies: If checked, CMM-Manager calibrates tips across all defined probe assemblies. If a probe rack is configured, probe assembly swaps occur automatically.
Opening the Tip Property dialog via View Detail... in Tip Manager displays all geometric, positional, and calibration parameters for a specific probe tip angle. When viewed via Tip Property, all fields are read-only so you can inspect the tip's parameters without accidentally modifying calibrated data.
This dialog can be used to check additional details about the specific tip that are not available on the main Tip Manager window such as the exact date and time stamp for the last calibration.
Selecting an Active Tip
To select an active probe tip for inspection, click on the specific tip entry (e.g., A0B0, A45B90) in the Tip Manager grid to highlight it then hit the Select button in the bottom-left area of the window. CMM-Manager will perform the tip change sequence and insert the Get Tip command into the current program.
Please ensure that the probe will not collide with an obstacle such as part and fixture before pressing the Select button.
Saving/Loading Tip File
A tip file (.tip file extension) is a database file that stores a collection of defined probe tip orientations and their associated calibration data. This is typically generated for 2 purposes.
- Backup and Restore: Saving your tip file to disk creates a backup file that can be reloaded to restore your probe setup.
- NOTE: The helpdesk file already contains tip information
- Commissioning an Identical System: If you purchased a CMM with identical setups, you could export a tip file from the existing machine then import it to the new one to match the settings.
CMM-Manager updates and stores tip calibration data automatically in the background, thus tip information will be retained even if the tip file is not saved before closing the application. Therefore, you do not strictly need to save a tip file manually after every calibration.
Loading the tip file will be grayed out for the Renishaw I++ system as tip and calibration data are stored in UCCserver.