CMM-Manager | Using Equator Gauging System
Programming with Renishaw Equator
Although the standard CMM-Manager knowledge applies, there are subtle differences that must be aware of when programming for Equator.
Requirements:
  1. CMM-Manager 2023 or newer with Renishaw Equator driver
    1. Standard Renishaw I++ driver will NOT be compatible
  2. The OEM controller is required for Equator
  3. A "golden part" for master and calibration files creation
    1. This part should be dimensionally close to the nominal as possible
  4. If the calibration file is intending to be used...
    1. CMM-Manager 2023 or newer (with the scanning option) should be installed on the CMM used for the calibration
    2. Although not required, styli used on the CMM and Equator should be similar as sharing the same program between 2 systems will be simpler
Info
Supported devices - CMM-Manager supports both the legacy Equator and the newer Equator-X systems as well as the Equator Auto Transfer System.  Learn More
Restrictions:
Following functions are not supported for Equator at the moment.
  1. 2D/3D Cloud Scan, CAD Curve Scan, CAD Cloud Scan
If the calibration file is being used, following operations are not supported due to the calibration file compatibility.
  1. Edge and Angle Points
  2. Relative Measure Function for 2D Features
  3. Hole Pattern
  4. If Pantec or Mitutoyo controller is used, following scan commands require re-pathing when switching between Equator and the CMM due to the capability differences.
    1. Pantec - Scanning ellipse and slot
    2. Mitutoyo - Scanning ellipse, line, slot, plane, cone, and sphere
Creating a Program:
For Equator to measure and compare data correctly, a reference frame must be established on the part prior to sending the comparison command. The picture below is accomplishing this by recalling the fixture reference frame, but the manual and/or DCC alignment can also be used.
Once the initial reference frame is established, insert the Compare command located under the Program tab in the Ribbon Bar. Ensure that Start and Auto Reset options are selected in the Action group. In the State group, select Master then enter the file path and name for the master file (.mst), which will be generated after performing the mastering with the golden part.
The file path here is referring to the location on the Renishaw PC, NOT the CMM-Manager workstation.
Now teach features for measurement purposes. Features taught after the Compare/Start command must be measured in the same reference frame for Equator to compensate correctly (i.e. no alignment function allowed until all necessary features are taught).
Add report commands once all features are programmed. Optionally, additional reference frames can be created for reporting purposes.
Any features measured prior to the Compare/Start command are not compensated by Equator thus should NOT be used for reporting.
Add another Compare command at the end of the program. Select End option under Action group.
Generating a Calibration File:
This step is necessary if you would like to see the Equator result to be tied to the absolute measurements from a CMM. Skip this section if Equator is being used as a "comparator".
Transfer the program as well as the "golden part" to a calibrated CMM. Double-click on Compare/Start in the Program Database. Select Calibrate under State group and specify the path and file name for the calibration file to be saved on the CMM workstation. The example shown below is simply saving it on a desktop.
If necessary, perform Align Part function and/or re-teach the fixture reference frame so that the program will adjust the path for the CMM.
Now execute the entire program to create a calibration file. For all scanned features, CMM-Manager will automatically multiply the point density by 10 while running in the calibration mode.
Once completed, transfer the calibration file to the Renishaw PC, re-open the program, and update the calibration file path to reflect the new location.
Generating a Master File:
The master file contains an inspection result from the golden part obtained by Equator. Subsequent measurement of the same part model will be compared against this data.
First, place the golden part on Equator and open the program.
Double-click on Compare\Start in the Program Database, select Master under Action group. Update the master file path and name if needed.
If necessary, perform Align Part and/or update the fixture reference frame so the program can start with appropriate path.
Run the entire program. At the end of the execution, the master file will be created at the specified name and location in the Renishaw PC.
Inspecting a Part:
Double-click on Compare/Start in the Program Database then switch the state to Measure. Check Use Calibration File option if you are intending to tie the measurement from the CMM.
Place the part in Equator and run the program. If only the master file is being used, deviations shown in the report is based on the last "golden part" used for mastering (i.e. zero deviation = dimensionally identical to the golden part).
The output will be an absolute measurement if the calibration file is utilized.
Re-Master & Re-Calibration:
Since Equator is intended to be used on the shop floor, it is advised to re-generate the master file to compensate for the environment change. Please refer to Renishaw's recommendation to determine the frequency.
Aside from the above reason, updating master and calibration files are recommended when following changes are made.
  1. Any changes to the program involving initial PCS, # of features, feature nominals, measurement order and path.
  2. The golden part has been altered or replaced.
Changing the Compare State via VB Variable:
A VBscript variable can be implemented to switch the state of Compare command without a manual intervention. In Compare/Start, select "Variable" state then specify the variable name. During the program execution, CMM-Manager will switch the state depending on the value stored in it (0 = Master, 1 = Measure, 2 = Calibrate). VBscript logic must be inserted prior to Compare/Start to assign the appropriate integer. Shop Floor Launcer can take an advantage of this option by calling the same part program but sending different VB value depending on the button pressed (Master Part 1 vs Measure Part 1).