During tool (tip) changes with an MCR20, the probe is normally inhibited between stalls by magnets in the MCR20 lids. During a "store" command during rack setup, an inhibit command must be sent to the controller to keep the probe inhibited after it moves away from the rack for to ask for the next tip. Not all Mitutoyo controllers will accept the inhibit command. There is no definitive list for this. It is know that "some" controllers with "some" firmware's work this way. The Probe menu > Get Probe Assembly command does work. Only the rack setup "store" command is affected on some controllers.
In order to "load" the rack using Get Probe Assembly:
- Mark all ports as "empty" in Rack setup. Since the ports are marked empty, any Get Probe Assembly will be treated as a manual change.
- Place the module that is to be stored in stall 1 of the rack onto your probe assembly.
- In Probe Assembly, select that tool from the list and click "Set Active".
- Now, simply do a Get Probe Assembly for each remaining tool in order. If the tips are correctly assigned in Probe Assembly, the current tip should be placed in the proper port, then the machine should ask for you to manually install the next tip. Repeat the process until all ports are full (re-select the first probe to drop off the last tool in the list).